Electrolyte Analysis
On this page
Direct answer
Electrolyte analysis measures sodium, potassium, chloride and bicarbonate, with reference ranges of 135-145, 3.5-5.0, 96-106 and 22-26 mEq/L respectively. Potassium tells the first lie in a haemolysed sample: a reported 7.1 mmol/L may be a healthy 4.2 because intracellular potassium far exceeds plasma. Most laboratories now use ion-selective electrodes rather than flame photometry, and pseudohyponatraemia is the main sodium trap.
What you must remember
- Reference ranges: sodium 135-145, potassium 3.5-5.0, chloride 96-106 and bicarbonate 22-26 mEq/L; total calcium 8.5-10.5 mg/dL and ionised calcium 1.1-1.3 mmol/L.
- Corrected calcium: measured calcium plus 0.8 times (4.0 minus serum albumin), since low albumin masks hypocalcaemia until corrected.
- Flame photometry: sodium emits at 589 nm and potassium at about 766 nm, ratioed against a lithium internal standard at 671 nm, a technique now largely historical.
- ISE modes: direct measurement on undiluted samples in blood gas analysers, and indirect measurement on diluted samples in chemistry analysers, where only the indirect mode is vulnerable to pseudohyponatraemia.
- Pseudohyperkalaemia: haemolysis, prolonged fist clenching under tourniquet and marked thrombocytosis or leucocytosis that releases potassium during clotting.
- Glucose correction: add about 1.6 mEq/L to measured sodium for every 100 mg/dL of glucose above normal, which matters in diabetic emergencies.
- Anion gap: sodium minus the sum of chloride and bicarbonate, normally 8-12 mEq/L, rising in ketoacidosis, lactic acidosis, renal failure and certain toxins.
Why measurement moved from flame to electrode
Flame photometry and ion-selective electrodes measure the same analytes by different physics, and the table shows what carried over and what changed.
| Feature | Flame photometry | Ion-selective electrode (ISE) |
|---|---|---|
| Principle | Light emitted by excited atoms in a flame | Membrane potential proportional to ion activity |
| Analytes | Sodium and potassium only | Sodium, potassium, chloride and bicarbonate |
| Reference | Lithium internal standard at 671 nm | Reference electrode against which the membrane potential is read |
| Wavelengths | Sodium 589 nm, potassium about 766 nm | Not applicable |
| Sample handling | Diluted and aspirated into a steady flame | Undiluted on blood gas analysers, diluted on chemistry analysers |
| Main hazard | Flame flicker and viscosity needing correction | Pseudohyponatraemia in indirect mode |
| Status | Historical, still in every syllabus | Routine on Indian benches |
The indirect ISE dilutes the sample and assumes a normal water fraction, and that assumption breaks in milky or myeloma plasma. This is why pseudohyponatraemia persisted into the modern era, and a direct ISE or blood gas analyser resolves it.
Where the marks are lost
Most marks are lost to sample artefacts and arithmetic rather than to reference ranges.
- Trap: Starting insulin and dextrose for a potassium of 7 before examining the plasma. Correct: Confirm that the sample is not haemolysed first, then give intravenous calcium gluconate for ECG changes and the insulin-dextrose shift.
- Trap: Ignoring glucose when reading sodium in diabetic ketoacidosis. Correct: Add about 1.6 mEq/L to measured sodium for every 100 mg/dL of glucose above normal, since the corrected value can be near normal.
- Trap: Adding potassium or subtracting sodium in the anion gap. Correct: The gap is sodium minus the sum of chloride and bicarbonate, normally 8-12 mEq/L.
Frequently asked questions
What are the reference ranges for the four core electrolytes?
The reference ranges are sodium 135-145, potassium 3.5-5.0, chloride 96-106 and bicarbonate 22-26 mEq/L. Bicarbonate shows some small variation between laboratory methods, so the reporting laboratory's own range should always be used.
What is the principle of flame photometry, and at which wavelengths?
Atoms excited in a flame emit characteristic light, with sodium at 589 nm and potassium at about 766 nm. Emission is compared with a lithium internal standard at 671 nm, and intensity is proportional to concentration.
What is pseudohyponatraemia and how is it resolved?
Pseudohyponatraemia occurs when marked hyperlipidaemia or hyperproteinaemia displaces plasma water, so indirect ISE reads sodium falsely low. Measuring with a direct ISE or a blood gas analyser resolves the falsely low reading.
What causes pseudohyperkalaemia in a laboratory sample?
Haemolysis, fist clenching during venepuncture and potassium released from markedly raised platelet or white cell counts during clotting all cause pseudohyperkalaemia. The sample should be checked before the reported value is treated.
How is sodium corrected for hyperglycaemia?
Add about 1.6 mEq/L of sodium for every 100 mg/dL of glucose above the normal range. The correction matters most in diabetic emergencies, where hyperglycaemia dilutes the measured sodium.
What is the anion gap and what raises it?
The anion gap is sodium minus the sum of chloride and bicarbonate, normally 8-12 mEq/L. Ketoacidosis, lactic acidosis, renal failure and certain toxins raise it.