Direct answer

Resolution, not magnification, defines a good microscope: a compound light microscope resolves detail only to about 0.2 micrometres, a limit set by the numerical aperture of the objective and the wavelength of visible light. Routine work uses 10x and 40x dry objectives plus the 100x oil immersion lens, whose oil matches the refractive index of glass. Beyond the light microscope, dark-ground illumination shows Treponema pallidum and electron microscopy resolves viruses.

What you must remember

  • Magnification: total magnification equals objective times eyepiece, so a 10x eyepiece with the 100x objective gives 1000x, the practical ceiling for light microscopy.
  • Numerical aperture: NA equals n times sin theta, and resolution improves as NA rises, reaching about 0.2 micrometres in air or oil.
  • Immersion oil: oil of refractive index about 1.515 prevents refraction losses at the glass-air boundary; use it only with the 100x objective, never letting the 40x lens touch an oiled slide.
  • Focusing: begin at low power, since microscopes are parfocal and higher objectives need only fine focus afterwards.
  • Dark ground and phase contrast: dark ground shows live, unstained Treponema pallidum in a chancre exudate, while phase contrast suits urine crystals, parasites and motility in wet mounts.
  • Fluorescence and electron microscopy: auramine O or labelled antibodies detect acid-fast bacilli, and electron microscopes, resolving around 0.2 nanometres, show viruses, with TEM for internal ultrastructure and SEM for surfaces.
  • Care: carry the microscope by the arm, clean oil with lens paper (xylene for dried oil), and cover it when idle.

From 10x to 100x without smashing a slide

A Leishman-stained peripheral smear is the practice slide for the focusing sequence examiners watch. The steps below follow it from low power to oil.

  1. Start at 10x: place the smear face up, centre a likely field and focus with the coarse knob until the film texture is clear.
  2. Move to 40x: swing the 40x objective in, use only a small fine-focus adjustment, and set the condenser and iris for a bright, unfaded image.
  3. Apply oil: raise the objectives, place one small drop of immersion oil on the area of interest and swing the 100x lens in.
  4. Lower the lens carefully: watching from the side, bring the lens down until it just touches the oil, then focus upward with the fine knob only.
  5. Assign each power a job: 10x judges cellularity and film quality, 40x surveys white cell distribution, and 100x carries the differential count, red cell morphology and platelet estimate.
  6. Clean up: wipe the oil from the lens with clean lens paper in one direction after use.

How the viva frames optics

Two confusions dominate optics questions, and both hinge on what limits the image.

  • Trap: believing higher total magnification always reveals more detail Correct: beyond the resolving limit, extra magnification only enlarges blur, which is empty magnification.
  • Trap: thinking oil sharpens every objective Correct: oil lifts the effective numerical aperture from about 1.0 in air to about 1.25, so it serves the 100x lens, where light would otherwise bend away and be lost.
  • Trap: assuming electron microscopy wins only through magnification Correct: it uses beams of far shorter wavelength, and resolution depends on wavelength as well as NA.

Frequently asked questions

What is the resolution limit of the light microscope and what sets it?

The limit is about 0.2 micrometres, fixed by the numerical aperture of the objective and the wavelength of visible light. Immersion oil raises the effective numerical aperture to about 1.25.

Which objective requires immersion oil and why?

The 100x objective needs immersion oil, because oil of refractive index about 1.515 prevents light loss at the slide-air interface and maximises resolution. Lower powers are used dry, and the 40x lens must never touch an oiled slide.

Which microscope shows Treponema pallidum in a chancre?

Dark-ground microscopy shows the thin, live spirochaete as a bright organism against a black background, without any stain. It is the standard method for this organism in a chancre exudate.

What are the applications of phase contrast microscopy?

Phase contrast suits unstained, live preparations such as urine crystals, protozoa, parasite eggs and bacterial motility in wet mounts. Staining would distort or kill these specimens.

How do TEM and SEM differ?

Transmission electron microscopy images thin sections for internal ultrastructure at up to nanometre resolution. Scanning electron microscopy coats and scans surfaces to give three-dimensional detail.

Why must focusing at 100x use only the fine adjustment knob?

The oil-immersion lens has almost no working distance, so the coarse knob drives the objective into the slide and cracks both. Only the fine knob moves the focal plane safely at this power.