UV-Visible Spectroscopy
On this page
Direct answer
Ultraviolet-visible spectroscopy measures the light a molecule absorbs as electrons move between energy levels, typically between 200 and 800 nanometres. Concentration follows from the Beer-Lambert law, A = εcl, which holds only where absorbance rises linearly with concentration. Chromophores confer absorption, and measurement is made at λmax because sensitivity is greatest there.
What you must remember
- Beer-Lambert law: A = εcl, with A = log(I0/I); it fails at high concentration, with turbidity or fluorescence, under polychromatic or stray light, and when the analyte associates or dissociates.
- Working rules: measure at λmax for maximum sensitivity and minimum wavelength error, and keep absorbance between about 0.2 and 0.8 where the calibration is linear.
- Chromophore catalogue: isolated alkene near 180-190 nm, carbonyl about 270-300 nm (weak), benzene about 254 nm; extended conjugation shifts absorption to longer wavelengths, so beta-carotene's polyene absorbs into the visible and looks orange.
- Shift vocabulary: bathochromic (longer wavelength), hypsochromic (shorter), hyperchromic (stronger) and hypochromic (weaker), the effects of auxochromes, conjugation and solvent on a band.
- Instrument anatomy: deuterium or hydrogen lamp for the ultraviolet and tungsten for the visible, a prism or grating monochromator, quartz cuvettes because glass absorbs below about 350 nm, and a photomultiplier or diode-array detector; double-beam designs cancel source and solvent variation.
- Solvent windows: water, ethanol and hexane are transparent through the near ultraviolet, each with a cut-off below which it absorbs.
- Applications: single-component tablet assays against reference standards, dissolution sampling and HPLC detection.
One tablet assay, step by step
A tablet assay shows how these rules combine in practice. The same logic applies to any single-component UV assay.
- Prepare the sample: twenty paracetamol tablets are weighed and powdered, and a portion equivalent to the average tablet weight is dissolved to a stock solution.
- Dilute to the linear range: an aliquot is diluted so that absorbance falls between about 0.2 and 0.8, because stray light and molecular crowding bend the Beer-Lambert line beyond it and the concentration becomes unreliable.
- Confirm λmax: a spectrum is scanned to confirm the maximum, which for paracetamol lies near 257 nm in alkaline medium, and the instrument is set there because the slope of absorbance against wavelength is zero at a maximum.
- Zero and read: the blank zeroes the double-beam instrument, then the standard and sample are read.
- Calculate by proportion: sample concentration equals standard concentration times the absorbance ratio, and the report carries the calculation and acceptance band.
Where the exam tries to catch you
Four traps recur on this topic.
- Trap: attributing every Beer's law deviation to high concentration. Correct: classify deviations as chemical (dissociation, association, reaction with solvent), instrumental (polychromatic and stray light) or physical (turbid solutions scattering light).
- Trap: swapping bathochromic and hyperchromic. Correct: a bathochromic shift moves the band's position to a longer wavelength, whereas a hyperchromic effect raises its height, which is intensity.
- Trap: choosing the lamp and cuvette casually. Correct: use quartz with the deuterium lamp for the ultraviolet, with glass acceptable only in the visible; measure at λmax for sensitivity and wavelength tolerance, not from habit.
- Trap: assuming UV spectrophotometry suits every assay. Correct: two absorbing drugs or interfering excipients push the laboratory toward HPLC, so knowing where the cheap method stops is part of knowing it.
Frequently asked questions
What is the Beer-Lambert law, and where does it fail?
The Beer-Lambert law states that absorbance equals molar absorptivity times path length times concentration (A = εcl). It fails at high concentration, with turbid or fluorescent samples, under polychromatic or stray light, and when the analyte changes chemically in solution.
What distinguishes a chromophore from an auxochrome?
A chromophore is the group responsible for absorption, such as C=C, C=O or an aromatic ring. An auxochrome is a substituent with lone pairs, such as –OH or –NH2, that cannot absorb alone but shifts and intensifies the chromophore's band.
Which shift does adding conjugation to a chromophore produce?
Extended conjugation lowers the energy gap between orbitals and moves absorption to longer wavelengths. This is a bathochromic (red) shift, usually with increased intensity, which is why polyenes such as beta-carotene approach the visible range.
Why is the analytical wavelength chosen at λmax?
At a maximum, the slope of absorbance against wavelength is zero, so small wavelength errors cause minimal absorbance error. Sensitivity is also greatest there, and quantitative work depends on that combination.
Which light sources and cuvettes serve the ultraviolet region?
A deuterium or hydrogen lamp provides continuous ultraviolet radiation and is paired with quartz cuvettes. Ordinary glass cuvettes are unsuitable below about 350 nanometres, while tungsten sources serve the visible region.