Angiography Suite Equipment

On this page
  1. Direct answer
  2. What you must remember
  3. Setting up a carotid angiogram run by run
  4. How the exam frames it
  5. Frequently asked questions
  6. Related topics

Direct answer

The angiography suite is built around a C-arm with an X-ray tube opposite a flat-panel detector, driven by an 80-100 kW generator for short high-milliampere runs and cine. Supporting kit includes a carbon-fibre floating tabletop, a power injector with flow and pressure limits, and a dose-area product meter. Biplane suites add a second C-arm for simultaneous projections, and rotational angiography builds three-dimensional vessel maps.

What you must remember

  • Generator and tube: 80-100 kW high-frequency generators, tube currents in the hundreds of milliamperes for digital runs, grid-controlled pulsed output, and fractional focal spots near 0.3-0.6 mm for magnification.
  • Detector: flat-panel caesium iodide detectors of 25 x 25 to 40 x 30 cm with selectable field sizes, and frame rates of 0.5-30 fps; magnification modes raise dose rate by the inverse of the area ratio.
  • Power injector: flow of about 4-6 mL/s for carotid and 15-25 mL/s for aortic runs, pressure limits of 300-900 psi matched to the catheter rating, a saline chaser, and contrast at 300-370 mg iodine/mL.
  • Table engineering: carbon-fibre top (low attenuation, radiolucent), floating and tilting motions, bariatric weight ratings, and longitudinal travel for runoff studies.
  • Safety architecture: leaded walls and control-room glass per AERB shielding design, ceiling-suspended lead-acrylic shields, table-side lead skirts, warning lights, interlocks and personnel monitoring.
  • Biplane and 3D: dual C-arms for neurointervention such as aneurysm coiling and thrombectomy, with rotational acquisition reconstructing 3D vascular maps.
  • Dose instrumentation: a DAP meter and cumulative entrance air kerma at the interventional reference point, displayed live and archived in procedure dose reports.

Setting up a carotid angiogram run by run

A diagnostic carotid study follows a fixed sequence, and each step protects both image quality and dose.

  1. Position the patient. The patient lies supine with the head secured on the radiolucent headrest, the femoral sheath in place, and a 5 F pigtail or selective catheter sitting in the aortic arch.

  2. Program the injector. Non-ionic contrast at about 300 mg iodine/mL runs at 15-18 mL/s for the arch injection, followed by a saline chase.

  3. Set the acquisition. Mask-mode digital subtraction runs at 3-6 frames per second, with a 0.3-0.6 mm focal spot if magnification is planned.

  4. Centre the anatomy. The table floats to centre the arch, and collimation tightens to the great vessels.

  5. Grab the mask. Mask frames are taken during a breath-hold before the injection fires.

When a run fails, the fix depends on the cause. A run blurred by swallowing needs a "swallow, don't breathe" instruction and perhaps a faster frame rate, while gas in the shoulder needs pixel-shifting of the mask or remasking. For selective internal carotid catheterisation, frame rates drop to 2-3 fps and injections become hand-shot 6-8 mL boluses, with road-maps guiding the wire.

How the exam frames it

Question papers return to the same design choices, and each has a precise answer.

  • Trap: saying the angiographic table is carbon fibre only for strength. Correct: carbon fibre is radiolucent, so images can be taken through the tabletop in any projection, and floating motion lets the operator track the catheter to the legs during runoff.
  • Trap: treating the power injector pressure limit as a minor setting. Correct: the limit protects the catheter, which can rupture if programmed pressure exceeds its rating, so injector and catheter settings must be matched.
  • Trap: describing the C-arm isocentre loosely. Correct: the isocentre is where the tube, detector and rotation axes intersect, and the anatomy of interest must sit there, especially in rotational acquisitions.

Frequently asked questions

Why is the angiographic table made of carbon fibre?

Carbon fibre attenuates very few X-rays while carrying bariatric weight, so images can be acquired through the tabletop in any projection. Its floating movements also allow rapid repositioning during runoff studies.

What parameters are programmed on a power injector?

Contrast volume, flow rate (roughly 4-6 mL/s for selective studies and 15-25 mL/s for aortic runs), injection rise time, a pressure limit matched to the catheter rating, and a saline chaser are set for each procedure.

What is the role of a biplane angiography system?

A biplane system has two independently moving C-arms that acquire anteroposterior and lateral projections simultaneously. This is the standard set-up for neurointerventional work such as aneurysm coiling and mechanical thrombectomy.

Which dose monitoring instruments must an interventional suite display?

An interventional suite displays a dose-area product meter and cumulative entrance air kerma at the interventional reference point. Both are shown live on the console and archived into the procedure dose report.

How does magnification affect patient dose in angiography?

Smaller detector field modes concentrate the same photon flux on less skin, which raises the entrance dose rate steeply. For this reason magnification is reserved for small critical regions and brief acquisitions.