Quality Assurance in Imaging
On this page
Direct answer
Quality assurance in imaging runs on four test tiers: acceptance testing at installation, commissioning to set baselines, status testing at periodic intervals, and constancy testing against those baselines. Core radiographic checks cover kVp accuracy within about 5%, output reproducibility, a half-value layer of at least 2.3 mm aluminium at 80 kVp and light-field congruence. AERB requires QA at installation and at defined intervals as a licence condition.
What you must remember
- Radiographic tolerances: kVp accuracy within 5%, timer accuracy about 5%, output coefficient of variation 0.05 or less, and output-mAs linearity within about 10%.
- Filtration: half-value layer at least 2.3 mm Al at 80 kVp, with total filtration of at least 2.5 mm Al above 70 kVp.
- Field and geometry: light field congruent with the radiation field within about 1-2% of source-to-image distance, beam perpendicularity, and focal spot checked with a star or slit tool.
- AEC testing: every chamber selection gives an exposure index within tolerance, the minimum response time is adequate, and the backup timer ends a failed termination.
- CT constancy: water-phantom CT number 0 ± 5 HU, uniformity and noise within baseline bands, laser and table-increment checks, and CTDIvol cross-checked with a pencil chamber.
- Mammography QC: phantom scoring of at least 4 fibres, 3 speck groups and 3 masses, a mean glandular dose of the order of 1-3 mGy per view, and compression force checks.
- Programme culture: repeat and reject analysis, comparison with national diagnostic reference levels, display monitor QC, and records ready for AERB inspection under the Radiation Safety Officer.
Auditing a radiographic room from morning to report
A room audit moves from the daily logbook to the monthly physics test, and every finding gets one of three verdicts.
- Check the logbook. Yesterday's constancy entries for kVp, output and light field must be signed and within baseline before the room opens.
- Run the daily digital test. A standard exposure over the same phantom is compared with the baseline exposure index, and detector calibration is updated if the vendor's tool flags drift.
- Run the monthly physics test. The physicist uses a kVp meter, an aluminium HVL set, a slit pattern for focal spot and a collimator tool, writing each result against the commissioning baseline. A kVp drifting beyond 5% is a finding, not a footnote.
- Review CT and mammography. The CT water phantom must read close to zero with noise inside tolerance, and the mammography image must show the fourth fibre and third speck group, because target quality comes before dose.
- Classify and sign. Each finding is marked within tolerance, investigate, or withdraw from service, and the Radiation Safety Officer signs the trend charts that AERB inspectors request by date.
How the exam frames it
Examiners ask which test catches which fault, so learn the mapping rather than the definitions.
- Trap: attributing noisy radiographs across all patients to positioning. Correct: they point to output or detector calibration faults, which a constancy test catches.
- Trap: attributing a contrast drift over weeks to processing alone. Correct: it suggests kVp or filtration drift, which the HVL test reveals.
- Trap: confusing quality control with quality assurance. Correct: control is the testing itself, while assurance is the whole managed programme, including training, procurement, reject analysis and audit.
- Trap: reading a low HVL as a safe margin. Correct: too little filtration lets soft photons through, which add skin dose and nothing to the image.
Frequently asked questions
What is the difference between acceptance and constancy testing?
Acceptance testing verifies a new installation against its specifications before any clinical use, while constancy testing checks performance frequently, from daily to monthly, against the commissioning baseline.
What HVL is required for a general radiographic unit at 80 kVp?
At least 2.3 mm of aluminium at 80 kVp, supported by total filtration of at least 2.5 mm Al above 70 kVp, so the soft photons that only add skin dose are filtered out.
Which daily CT quality checks are performed on a water phantom?
The water CT number must read within 0 ± 5 HU, uniformity across the phantom and image noise must stay within the baseline band, and laser alignment and table increment checks are included.
What phantom score defines acceptable mammography image quality?
At least 4 fibres, 3 speck groups and 3 masses must be visible in the phantom image, with mean glandular dose held to a few milligray per view.
When must full quality assurance be repeated?
After major repairs, tube or detector replacement, relocation or software changes that affect image quality or dose, and at the defined periodic intervals in between.