Ventilator Breath Types, Triggering and ARDSNet Protocol
On this page
Direct answer
Ventilator breath types are defined by three answers: what triggers the breath, what limits it and what cycles it off. The ARDSNet protocol anchors every setting in lung-protective ventilation: tidal volume of 6 mL/kg of predicted body weight, plateau pressure at or below 30 cm H2O and driving pressure kept low. Everything else in ventilator management is titration and troubleshooting around that scaffold.
What you must remember
- Predicted body weight: male 50 kg plus 0.91 kg per cm over 152 cm; female 45.5 kg plus the same 0.91 kg per cm. Tidal volume is 6 mL/kg (range 4-8) of this weight, never the measured weight.
- Starting rate: 12-20 breaths per minute, higher in ARDS and lower in obstruction; adjust on pH and CO2 and watch for auto-PEEP when expiration is incomplete.
- PEEP floor: 5 cm H2O is the standard minimum, recruiting alveoli and preventing atelectrauma; higher values follow ARDS tables or oesophageal manometry.
- Plateau and driving pressure: plateau, read during an inspiratory hold at zero flow, stays at or below 30 cm H2O; driving pressure (plateau minus PEEP) is best kept under 15.
- Peak versus plateau: peak reflects airway resistance and plateau reflects compliance, so a high peak with a normal plateau points to secretions, kinking, biting or bronchospasm rather than a stiff lung.
- Trigger and cycle: pressure or flow sensing of patient effort triggers the breath; it cycles off on time in pressure control, on volume in volume control, and on flow at about 25% of peak inspiratory flow in pressure support.
- I:E ratio: starts at 1:2; obstructed patients need 1:3 or 1:4 with lower rates to empty the lung, accepting permissive hypercapnia.
- Complications: ventilator-induced lung injury (barotrauma, volutrauma, atelectrauma, biotrauma), ventilator-associated pneumonia (head-up 30-45 degrees, oral care, subglottic suction where available) and reduced venous return at high intrathoracic pressure.
Settings for one ARDS patient
Take a 35-year-old woman, 165 cm tall, ventilated for ARDS from pneumonia, and check each setting against its rule.
| Setting | Rule | Her value |
|---|---|---|
| Predicted weight | 45.5 kg female plus 0.91 kg per cm over 152 cm | About 57 kg |
| Tidal volume | 6 mL/kg of predicted weight | About 340 mL |
| Rate | 12-20 per minute, higher in ARDS | 24 |
| PEEP | At least 5 cm H2O, higher per ARDS tables | 10 |
| FiO2 | Wean to keep SpO2 88-95% | Started high, then weaned |
| Plateau pressure | At or below 30 cm H2O | 28 |
| Driving pressure | Plateau minus PEEP, best under 15 | 18 |
The plateau passes, but a driving pressure of 18 is above 15, so the volume comes down further and the rate goes up, tolerating a higher CO2 in exchange for a protected lung. Predicted weight anchors the whole calculation, so a height error moves every later volume.
Where the exam tries to catch you
Three errors recur in both viva answers and bedside practice.
- Trap: using actual weight for tidal volume Correct: a 120 kg obese man has the lungs of a 60 kg man, so 6 mL/kg of measured weight volutraumatises him.
- Trap: confusing peak with plateau pressure Correct: one inspiratory hold separates a tube problem from a lung problem in seconds, and candidates who describe the manoeuvre earn the viva.
- Trap: chasing oxygenation with FiO2 alone, or chasing normocapnia in ARDS Correct: recruitable lung needs PEEP and permissive hypercapnia is the protective strategy. SpO2 and PaO2 report oxygenation, while ETCO2 and PaCO2 report ventilation.
Frequently asked questions
Why is predicted body weight used for tidal volume rather than actual weight?
Lung size tracks height and sex rather than body mass, so the excess weight in obesity is not lung. Using actual weight for the 6 mL/kg rule gives volumes that overdistend the lungs.
How is the tidal volume calculated for a 170 cm ventilated woman?
For a 170 cm ventilated woman, predicted weight is 45.5 plus 0.91 times 18, the centimetres above 152, which is about 62 kg. At 6 mL/kg of that predicted weight, the tidal volume is roughly 370 mL, not a figure based on her actual weight.
What is the plateau pressure limit and how is it measured?
Plateau pressure should stay at or below 30 cm H2O, read during an inspiratory hold at zero flow. It reflects alveolar pressure rather than airway resistance, which is why the hold is used to judge the lung itself.
How do peak and plateau pressures differ?
Peak pressure includes airway resistance, while plateau pressure excludes it. A high peak with a normal plateau points to resistance from secretions, kinking or biting, whereas a high peak and a high plateau together indicate poor compliance.
What should you do for a high-pressure alarm on a ventilator?
Disconnect the patient and hand-ventilate with 100% oxygen through a self-inflating bag, a manoeuvre that is both diagnostic and therapeutic. Then look, listen and feel: bilateral air entry with improving compliance points to secretions or biting, while absent left-sided sounds with hypotension and tracheal shift mean tension pneumothorax, which needs needle decompression and then a chest drain.