Cone Beam CT
On this page
Direct answer
Cone beam CT acquires a whole volume in one rotation: a cone-shaped X-ray beam strikes a flat-panel detector as the tube turns through 180 to 360 degrees in 5 to 40 seconds. A Feldkamp reconstruction converts the projections into isotropic voxels as fine as 0.08 to 0.4 mm. The technique suits small regions such as dentoalveolar imaging, and dental effective doses run from tens to a few hundred microsieverts.
What you must remember
- Acquisition geometry: one 180-360 degree rotation with a flat-panel detector, typically 15 x 15 cm to 30 x 40 cm; one rotation gives one volume.
- Voxel size: 0.08-0.4 mm gives near-isotropic resolution, superior to fan-beam CT for small bony detail.
- Field of view: small (under 5 cm, one jaw quadrant), medium (5-10 cm, both jaws) and large (over 10 cm, skull base to hyoid); dose scales with the imaged volume, so the field must match the question.
- Reconstruction: Feldkamp back-projection; circular trajectories leave artefacts at the scan extremes, one reason long Z-axis coverage reduces accuracy.
- Hounsfield reliability: CBCT numbers are approximate and not water-calibrated, so nodule density or enhancement must never be reported from CBCT.
- Artefact profile: metal streaks from amalgam, ring artefacts from faulty detector rows, cupping from beam hardening in dense bone, scatter streaks, and motion blur over a 20-40 second exposure.
- IGRT role: kilovoltage CBCT on the linac verifies patient position against the planning CT immediately before treatment.
- Indian regulation: CBCT units fall under AERB requirements, with registration through the e-LORA system and quality assurance at acceptance and periodic intervals.
Planning an implant with cone beam CT
A 26-year-old needs implant planning in the mandibular molar region. A panoramic image cannot show buccolingual width or the canal's three-dimensional course, so small-field cone beam CT is chosen, and the steps below follow the scan from start to measurement.
- Choose the smallest useful field: a 5 x 5 cm small field covering the implant site plus a margin, because dose follows the irradiated volume.
- Select the finest voxel: the smallest setting the unit allows, 0.08-0.125 mm for dentoalveolar work.
- Orient the arc: keep the region of interest at the rotation centre, where resolution is best.
- Immobilise the patient: the patient bites gently and stays still for the 15-20 second exposure, since motion blurs the volume.
- Generate the reformats: cross-sectional, panoramic and 3D views are all produced from the one dataset.
- Measure the critical relation: trace the inferior alveolar canal and measure its buccolingual relation to the planned implant directly.
Where the exam tries to catch you
Cone beam CT questions usually test judgement about dose, accuracy and whether the modality is the right tool at all.
- Trap: calling CBCT "3D radiography with no dose concern" Correct: a large-field scan's effective dose approaches that of a low-dose MDCT, so justification and the smallest field still apply.
- Trap: reading CBCT Hounsfield numbers as tissue density Correct: the numbers are approximate and not water-calibrated, so density or enhancement cannot be quantified.
- Trap: calling a faulty detector row a streak Correct: it produces a ring centred on the rotation axis, because the same pixel traces a circle through every projection.
- Trap: choosing CBCT for soft-tissue detail Correct: it is soft-tissue-poor, so multidetector CT is preferred when soft-tissue contrast matters.
Frequently asked questions
How does cone beam CT differ from multidetector CT in acquisition?
Cone beam CT captures the whole volume in one rotation on a flat panel, while multidetector CT irradiates thin fan-beam slices as the table feeds. Cone beam CT wins on resolution and dose for small fields, whereas multidetector CT gives better soft-tissue contrast and Hounsfield accuracy.
Why are Hounsfield units from cone beam CT unreliable?
Wide cone beams generate heavy scatter, and most small-field units have no anti-scatter grid. Cone beam CT also lacks water-based calibration, so its CT numbers are approximate and should not be used for density quantification.
Which artefact indicates a defective detector row in cone beam CT?
A ring artefact centred on the rotation axis, because the faulty pixel traces a circle through every projection. Streaks point instead to metal or photon starvation, so describing a detector fault as a streak loses the point.
What field of view should be selected for single-implant planning?
A small field of about 5 cm or less, confined to the implant region, because scan dose scales with the irradiated volume. The smallest field that still answers the clinical question is the correct choice.
What is the role of cone beam CT in image-guided radiotherapy?
Kilovoltage cone beam CT on the linac produces a low-dose 3D image of the patient in the treatment position. That image is matched to the planning CT to compute and apply daily couch corrections before each fraction.